A crocodile clip probe is used to measure a digital signal. The input to the scope is a 1 Meg resistance in parallel with 20 pF. Furthermore, the probe is 0.8 m long and for the co-ax C = 100pF/m. A 10 ns rise time (10% - 90%) pulse is measured. The source impedance for the digital signal output being measured is 60 ohms. At the high frequencies (around the signal bandwidth), what is the typical error made (in relationship to the signal) in the measurements at these high end frequencies? Enter the value of the error in percentage without the %.

Question

A crocodile clip probe is used to measure a digital signal. The input to the scope is a 1 Meg resistance in parallel with 20 pF. Furthermore, the probe is 0.8 m long and for the co-ax C = 100pF/m. A 10 ns rise time (10% - 90%) pulse is measured. The source impedance for the digital signal output being measured is 60 ohms. At the high frequencies (around the signal bandwidth), what is the typical error made (in relationship to the signal) in the measurements at these high end frequencies? Enter the value of the error in percentage without the %.

...expand
🧐 Not the exact question you are looking for?Go ask a question

Solution

To calculate the error made in the measurements at high frequencies, we first need to calculate the total capacitance and the total resistance of the system.

  1. Calculate the total capacitance:

The total capacitance (C_total) is the sum of the capacitance of the probe (C_probe) and the capacitance of the scope (C_scope).

C_probe = length of the probe * capacitance per meter = 0.8 m * 100 pF/m = 80 pF

C_scope = 20 pF

So, C_total = C_probe + C_scope = 80 pF + 20 pF = 100 pF

  1. Calculate the total resistance:

The total resistance (R_total) is the sum of the resistance of the probe (R_probe) and the resistance of the scope (R_scope).

R_probe = 60 ohms (source impedance for the digital signal output)

R_scope = 1 Meg ohm = 1,000,000 ohms

So, R_total = R_probe + R_scope = 60 ohms + 1,000,000 ohms = 1,000,060 ohms

  1. Calculate the time constant (τ):

The time constant of a system is the product of the total resistance and the total capacitance.

τ = R_total * C_total = 1,000,060 ohms * 100 pF = 100 ns

  1. Calculate the error:

The error is the difference between the rise time of the pulse and the time constant, divided by the rise time of the pulse, expressed as a percentage.

Error = ((τ - rise time) / rise time) * 100% = ((100 ns - 10 ns) / 10 ns) * 100% = 900%

So, the typical error made in the measurements at these high end frequencies is 900%.

This problem has been solved

Similar Questions

A digital signal comes from a circuit with 1 ohm source impedance. You probe the signal. The probe presents as 1 ohm over the entire bandwidth of the mesurement setup. What percentage error will you make in the voltage measurement? Enter the answer in % without any signs.

You measure a 1 ns risetime digital signal. What frequency wil you use to check if the pF's in the probe/scope/cable combination are going to cause probe loading? Enter the answer in MHz without the unit.

51. A student found out that x-radiation has an ENERGY of 4.60 x 10^-14 J in the experiment. The THEORETICAL WAVELENGTH of x-radiation is listed as 4.10 x 10^-12 m. What is the percent error of ENERGY in this experiment? Report ONLY NUMERIC ANSWER (with no percent sign)  in the correct sig figs. You may use Planck's constant value = 6.626 x 10^-34 J-s, and speed of light = 3.00 x 10^8 m/s1 pointYour answer

Two resistance are given as R1=(10±0.5) Ω and R2=(15±0.5) Ω. The percentage error in the measurement of equivalent resistance when they are connected in parallel is

how to find percentage error

1/1

Upgrade your grade with Knowee

Get personalized homework help. Review tough concepts in more detail, or go deeper into your topic by exploring other relevant questions.